    Update on #40C Sensor Shipments

    For some time now we have been studying and checking NRG #40 anemometer tolerances and manufacturing processes in tremendous detail to understand what has been causing some of the sensors to intermittently underreport wind speed. 

     Recently we reported that a vibratory mode had been identified as the cause of the slow down. As a result of our analysis and testing, we have revised the specifications of certain materials and tightened the tolerances of certain dimensions of the sensor.  These improvements have eliminated the vibratory mode in test sensors in our laboratory. We have further validated these improvements in wind tunnel tests and in limited field tests. Final validation of these improvements must await the results of longer term field data which will take another 6 to eight weeks.

    We are confident, based upon the results of our laboratory testing, that these changes have a high potential to significantly reduce the probability that a sensor will degrade in the field. Therefore, we believe that it is appropriate to integrate these modifications into the NRG#40 immediately.  As of January 5, 2009, all sensors will ship with the improvements.

    Larry Jacobs
    NRG Systems Inc.
    Senior Marketing and Sales Engineer
    Filed under: ,
  • 03-10-2009 5:03 In reply to

    Re: Update on #40C Sensor Shipments

     

  • 03-10-2009 5:22 In reply to

    Re: Update on #40C Sensor Shipments

    To the NRG Technical Support,

    Our company, Misibis Resort and Management, Inc. BOUGHT TWO UNITS of NRG-NOW System 50m sets of wind mast in July 2007, and another TWO units in June 2008. these units  have not yet been installed until our consultant, GARRAD HASSAN, suggested that # 40c Anemometer has technical problem.

    In view of the above, can you replace our anemometers without costs on our part. Request your advise on this matter.

    please email me at mccaliva@sunwet.com.ph 

    thanks and regards.

    mar caliva

    Philippines 

     

     

     

  • 03-10-2009 5:28 In reply to

    Re: Update on #40C Sensor Shipments

    To the NRG Technical Support,

    Erratum. Sorry my email ad is mccaliva@sunwest.com.ph.

    Thanks.

    Mar Caliva

    Philippines

  • 03-11-2009 14:41 In reply to

    Re: Update on #40C Sensor Shipments

     I have forwarded your request to our sales staff so that they can arrange to send you some replacement sensors.

  • 03-31-2009 13:27 In reply to

    Re: Update on #40C Sensor Shipments

     Hi,

     We bought 10 kits of anemometers from a sale representant in Brazil. Five (5) towers already were installed, but there are others five (5) not installed yet.

    Can you replace this anemometers not used yet? Are there costs of shipment to do this?

    Erick Castro
    ecastro@endesabr.com.br
  • 03-31-2009 15:42 In reply to

    Re: Update on #40C Sensor Shipments

     We would be happy to replace any anemometers still under warranty free of charge (free shipping, too).  Please contact your NRG salesperson for assistance.

    Filed under:
  • 04-01-2009 11:28 In reply to

    Replacement of #40C

     My kits contains #40C. The replacement will be for a updated #40C or for #40H?

    Otherwhise, Is the #40H a update for #40C or just other anemometer kind?

     Thanks,

    Erick.

    Erick Castro
    ecastro@endesabr.com.br
  • 04-01-2009 11:43 In reply to

    Re: Replacement of #40C

    Hello, Erick - The #40H is NOT the same sensor as the #40C.  The #40H Hall Effect anemometer requires power and provides a square wave output signal.  The #40C provides a self-generated AC sine wave output signal.

    Filed under: ,
  • 04-01-2009 12:41 In reply to

    Re: Replacement of #40C

    Thanks Julia,

    So I was thinking correctly.

    My point now is: Does #40H sensor have presented vibratory mode problem too?

    Erick Castro
    ecastro@endesabr.com.br
  • 04-01-2009 13:33 In reply to

    Re: Replacement of #40C

    #40H anemometers are also potentially affected by the vibratory mode problem.

    Filed under: ,