Jump to content

Symphonie Plus 3 Testing


Recommended Posts

Dear Sirs, i just wanted to know if there is any method to test the accuracy of Symphonie Plus 3 data logger at site i.e. in case we want to check if the logger is properly recording (converting signals from anemometers / wind vane) the correct values of wind speed etc.

 

A similar sensor testing is mentioned in application note. http://www.nrgsystems.com/TechSupport/KnowledgeBase.aspx?id=4&category=16

 

Regards,

Zeeshan

Link to comment
Share on other sites

 Hi Zeeshan,

Good question. It is possible to test the frequency inputs:

  • program the channel in the logger with these values
    • scale factor = 1
    • offset = 0.
  • Input a squarewave of known frequency 
Whatever the frequency of the input should match the values recorded in the datafile.

 The analog channels can be tested in a similar manner, though the data will be scaled according to the number of steps in the input voltage range.

To test the analogs,

  • Use a 3153 Temperature SCM. This SCM is setup for a full scale voltage reading of 2.500 volts.
  • As it is a 10-bid a/d converter, program the channel in the logger with these values
    • scale factor = 0.0024
    • offset = 0.
  • Then, apply voltages from 0 - 2.500 V and verify that the datafile matches your inputted voltage.

For the vane inputs you can apply a voltage from 0 - 2.500 V using the default scale factor and offset. The scaled readings on the vane channels will sweep from 4 - 356 degrees in a linear fashion.

 Note that you can view the logger's display to verify the output, but there may be a slight discrepancy between the inputted voltage and displayed value. The data in the RWD file will be precise.

Dave

Link to comment
Share on other sites

Archived

This topic is now archived and is closed to further replies.

×
×
  • Create New...