I am using SR11-05 sensor in the first

The information of sensor:

Calibration result: - sensittivity: S= 20.67 x 10^-6 V /(W/m2) ; Calibration uncertainty +- 0.28 x 10^-6 V /(W/m2)

In this case, Can i use the formula " scale factor = 119.21 / Calibration sheet number" ? 

(S = 119.21 / 20.67 = 5.76; offset=0)

 

Please help me

 

Thank you very much

 

vinhquyen

Share this post


Link to post
Share on other sites

Hello Vinhquyen,

 

Thank you for your post.

 

To properly scale thermopile sensors you need to consult the "Scaling_Thermopile_Sensors_AppNote.pdf". I'm not attaching it here because the information may change overtime and post may not be updated with the most recent version. Please refer to our website or email support@rnrgsystems.com for a current copy.

 

With a Hukseflux SR11, you'll need to use a "SCM Constant" of 29.875 and divide by the sensors calibration (sensitivity) value.

 

Scale Factor = SCM Constant / Sensitivity.

 

Scale Factor = 29.875 / 20.67 = 1.445.

 

The offset will be 0.

 

Please contact support@rnrgsystems.com with any further questions.

 

Thank you!

 

Pearse

 

 

Share this post


Link to post
Share on other sites

Join the conversation

You are posting as a guest. If you have an account, sign in now to post with your account.
Note: Your post will require moderator approval before it will be visible.

Guest
Reply to this topic...

×   Pasted as rich text.   Paste as plain text instead

  Only 75 emoji are allowed.

×   Your link has been automatically embedded.   Display as a link instead

×   Your previous content has been restored.   Clear editor

×   You cannot paste images directly. Upload or insert images from URL.

Loading...

×
×
  • Create New...